All 334 of the Biophysics quiz, by syllabus topic. Open an answer for the explanation; Practise opens the question in the quiz, or press Show all answers.
Subject quizzes: written for this site from the syllabus topics; every answer comes with its explanation.
Readings that agree closely with each other but all lie far from the true value are:
Aaccurate but not precise
Bprecise but not accurate
Cboth precise and accurate
Dneither precise nor accurate
Answer
Answer B. precise but not accurate
Precision is how closely repeated readings agree; accuracy is how close they are to the true value. A tight cluster in the wrong place points to a systematic error.
A number written down wrongly or a scale read wrongly is a:
Asystematic error
Bgross error (a mistake)
Crelative error
Drandom error
Answer
Answer B. gross error (a mistake)
A gross error is a one-off blunder; such a reading is found as an outlier and discarded, not averaged in. Systematic errors repeat in the same direction; random errors scatter both ways.
Which SI base unit is used for the amount of substance?
Athe candela
Bthe kelvin
Cthe mole
Dthe ampere
Answer
Answer C. the mole
Amount of substance is one of the seven SI base quantities and its base unit is the mole; the kelvin measures thermodynamic temperature, the candela luminous intensity and the ampere electric current.
The pascal, the SI unit of pressure, is equivalent to which combination of other SI units?
Aone newton per metre
Bone newton per square metre
Cone newton metre
Done joule per second
Answer
Answer B. one newton per square metre
Pressure is force divided by area, so 1 Pa = 1 N/m²; a newton metre is a joule of work, a joule per second is a watt of power, and a newton per metre is a surface tension.
A clinical thermometer reads 1.5 °C too high at every temperature. Which step will improve the accuracy of its future readings?
Ataking ten readings and averaging them
Bdiscarding the highest and the lowest reading
Ccalibrating it against a standard thermometer
Dreading the scale from directly in front of it
Answer
Answer C. calibrating it against a standard thermometer
A constant offset is a systematic error: it shifts every reading the same way, so repeating and averaging cannot remove it, and only calibration against a standard reveals and corrects it.
In analytical measurement, what does the limit of detection of a method mean?
Athe largest quantity it can measure without saturating
Bthe smallest quantity it can tell apart from a blank
Cthe difference between its reading and the true value
Dthe smallest division marked on its scale
Answer
Answer B. the smallest quantity it can tell apart from a blank
The limit of detection is the smallest amount whose signal is reliably separable from that of a blank; the smallest scale division is the instrument's resolution, and the gap from the true value is its error.
A vernier calliper has main-scale divisions of 1 mm and a vernier scale of 10 divisions spanning 9 mm. Its least count is:
A0.01 mm
B0.1 mm
C0.5 mm
D1 mm
Answer
Answer B. 0.1 mm
The least count is one main-scale division minus one vernier division: 1 mm − 0.9 mm = 0.1 mm, which is the resolution of the instrument: the smallest difference in length it can register.
Two quantities whose relative errors are 3% and 4% are multiplied together. If the errors are independent and random, so that they combine in quadrature, the relative error of the product is:
A1%
B3.5%
C5%
D7%
Answer
Answer C. 5%
For a product the relative errors combine, and independent random errors add in quadrature: √(3² + 4²) = 5%. Simply adding them, the maximum-error rule, would give the wider 7% bound.
In a set of repeated readings of the same quantity, what does the standard deviation describe?
Athe spread of the single readings about their mean
Bthe difference between the mean and the true value
Cthe largest difference between any two readings
Dhow precisely the mean of the set is known
Answer
Answer A. the spread of the single readings about their mean
The standard deviation measures the scatter of individual readings; how well the mean itself is known is the standard error of the mean, SD/√n, and the gap between the mean and the true value is a matter of accuracy.
A spring balance reads 0.2 N with nothing hanging from it. Every load weighed on it without resetting the pointer is recorded as:
A0.2 N too high, a systematic zero error
B0.2 N too low, a systematic zero error
Cscattered at random about the true value
Dcorrect, because the scale is linear
Answer
Answer A. 0.2 N too high, a systematic zero error
A zero offset adds the same 0.2 N to every reading, so all results are shifted the same way; this is a systematic error that repeating will not reduce, and it is cured by resetting the pointer or subtracting the zero reading.
A least-squares fit to a set of measurements gives a correlation coefficient of −1. What does that value say about the points?
Athey show no relation to one another
Bthey lie on a curve that falls ever more steeply
Cthey lie exactly on a straight line that rises
Dthey lie exactly on a straight line that falls
Answer
Answer D. they lie exactly on a straight line that falls
A correlation coefficient of −1 means a perfect straight-line relation in which y falls as x rises; +1 would be a perfect rising line and 0 no linear relation at all, while the least-squares line itself minimises the squared vertical distances.
Averaging 9 independent readings instead of taking a single reading reduces the random uncertainty of the result by a factor of:
A1.5
B81
C9
D3
Answer
Answer D. 3
The standard error of the mean is SD/√n, so with n = 9 the uncertainty shrinks by √9 = 3; averaging does nothing at all to a systematic error, however many readings are taken.
In laminar flow through a tube, the velocity profile is:
Aparabolic: fastest at the centre, zero at the wall
Bflat: the same speed everywhere
Cfastest at the wall, zero at the centre
Drandom from moment to moment
Answer
Answer A. parabolic: fastest at the centre, zero at the wall
Layers slide over each other; the layer touching the wall is held still and the speed rises to a maximum on the axis. A random, swirling pattern is turbulence.
For a spherical alveolus, Laplace's law gives the pressure inside as:
Ap = T / (4r)
Bp = T / r
Cp = T / (2r)
Dp = 2T / r
Answer
Answer D. p = 2T / r
A sphere has two curvatures, so p = 2T/r: at the same surface tension a smaller alveolus has a higher pressure, which surfactant offsets by lowering the tension.
An adult stands upright, so that a vein in the foot lies 1.2 m inferior to the heart. With blood density 1050 kg/m³ and g = 9.81 m/s², the extra hydrostatic pressure in that vein is about:
A1.2 kPa
B4 kPa
C12 kPa
D100 kPa
Answer
Answer C. 12 kPa
The hydrostatic term ρgh = 1050 × 9.81 × 1.2 ≈ 12.4 kPa, roughly 93 mmHg added to the venous pressure; the venous valves and the muscle pump work against it to return blood from the foot.
In which part of the systemic circulation is the Reynolds number of the blood flow the highest?
Aa venule
Ban arteriole
Ca capillary
Dthe aorta
Answer
Answer D. the aorta
Re = ρvd/η rises with both speed and diameter, and both are greatest in the aorta, where flow can briefly turn turbulent during systole; in a capillary the tiny diameter and the slow flow give a Reynolds number far under 1.
By Poiseuille's law, tripling the length of a rigid tube while the pressure difference across it is unchanged alters the flow to:
Aone third
Bthree times as great
Cone ninth
Done eighty-first
Answer
Answer A. one third
Laminar flow Q = πr⁴Δp/(8ηl) is inversely proportional to the length, so three times the length leaves one third of the flow; it is the radius, not the length, that enters as a fourth power.
Three vessels, each of resistance R, are connected in series. Their combined resistance is:
A3R
BR
C9R
DR/3
Answer
Answer A. 3R
Resistances in series add, giving R + R + R = 3R; it is parallel vessels whose reciprocals add, which is why the vast parallel capillary bed offers far less resistance than one capillary does.
What produces the Korotkoff sounds heard through a stethoscope while the cuff pressure falls during a blood-pressure measurement?
Avibration of the rubber cuff against the skin
Bthe closing of the aortic and the mitral valve
Cturbulence in the partly compressed artery
Dlaminar flow returning to the open artery
Answer
Answer C. turbulence in the partly compressed artery
Once the cuff no longer occludes the artery completely, blood is driven through the narrowed segment fast enough to exceed a Reynolds number of about 2000, and the turbulence is audible; fully laminar flow is silent.
During a blood-pressure measurement, the cuff pressure at which the Korotkoff sounds first become audible is taken as the:
Amean arterial pressure
Bdiastolic pressure
Csystolic pressure
Dpulse pressure
Answer
Answer C. systolic pressure
Sounds begin when the peak arterial pressure just exceeds the cuff pressure and squirts blood through the narrowed artery, so that reading is the systolic pressure; they fade at the diastolic pressure, when flow runs on laminar and quiet.
The apparent viscosity of blood flowing along vessels narrower than about 0.3 mm, compared with its viscosity in wide tubes, is:
Alower, the Fåhræus-Lindqvist effect
Bthe same, since viscosity is a property of the fluid
Chigher, but only at low shear rates
Dhigher, because drag at the wall dominates
Answer
Answer A. lower, the Fåhræus-Lindqvist effect
Blood is non-Newtonian: in narrow vessels the red cells stream along the axis and leave a cell-poor plasma layer at the wall, so the apparent viscosity falls, which is the Fåhræus-Lindqvist effect.
In steady laminar flow along a cylindrical vessel, where is the shear rate the greatest?
Aat the vessel wall
Bmidway between the centre and the wall
Cat the centre of the lumen
Dthe same everywhere across the lumen
Answer
Answer A. at the vessel wall
The laminar profile is parabolic, fastest on the axis and zero at the wall, so the velocity gradient, which is the shear rate, is steepest at the wall and vanishes at the centre; the endothelium therefore feels the highest shear stress.
The speed at which the arterial pulse wave travels along the aorta rises when the aortic wall becomes:
Athinner
Bmore compliant
Cstiffer
Dwider
Answer
Answer C. stiffer
Pulse wave velocity grows with the elastic modulus of the wall, so a stiff, poorly compliant aorta carries the pressure wave faster, while a compliant, wider or thinner-walled vessel slows it; viscosity of the blood does not enter.
Compared with water or soft tissue, air absorbs ultrasound:
Aless, at every frequency
Bequally
Conly below 20 kHz
Dfar more, and more at higher frequencies
Answer
Answer D. far more, and more at higher frequencies
Air absorbs ultrasound strongly, and more at higher frequencies: the absorption lab measures this. Gel under a probe is needed mainly for another reason: at an air gap the impedance mismatch reflects over 99.9% of the beam.
Ultrasound cannot image through lung or bowel gas because:
Anearly all the energy is reflected at the tissue-gas boundary
Bgas absorbs no ultrasound
Cthe speed of sound in gas is too high
Dgas is ionised by the beam
Answer
Answer A. nearly all the energy is reflected at the tissue-gas boundary
Air's acoustic impedance is thousands of times lower than tissue's, so almost all the beam reflects at the boundary and nothing returns from the far side.
Dthe movement of reflectors along one line against time
Answer
Answer D. the movement of reflectors along one line against time
M-mode follows one scan line over time, ideal for valves and the ventricular wall; A-mode plots amplitude against depth and B-mode makes the 2-D image.
Bits density multiplied by the speed of sound in it
Cits density multiplied by the frequency of the wave
Dthe speed of sound in it divided by the wavelength
Answer
Answer B. its density multiplied by the speed of sound in it
Acoustic impedance Z = density × speed, and echoes form wherever Z changes between two tissues; the enormous mismatch at a tissue-air boundary is why nearly the whole beam is reflected there.
Diagnostic ultrasound imaging of the body uses frequencies in the range:
Aabout 20 to 200 kHz
Babout 200 to 800 kHz
Cabout 2 to 15 MHz
Dabout 50 to 300 MHz
Answer
Answer C. about 2 to 15 MHz
Clinical probes work between roughly 2 and 15 MHz, well over the 20 kHz threshold of ultrasound: high enough for millimetre detail, since the wavelength in soft tissue is about 0.3 mm at 5 MHz, yet low enough to penetrate.
A pulse-echo scanner turns the echo delay t into a depth using d = ct/2. Why is the factor of two needed?
Athe echo is reflected twice at the boundary
Btwo crystals are used, one sending and one receiving
Cthe pulse travels to the reflector and back again
Dsound travels in tissue at half its speed in water
Answer
Answer C. the pulse travels to the reflector and back again
The measured time covers the round trip, so the depth of the reflector is half of c × t; the speed in soft tissue, about 1540 m/s, is close to that in water rather than half of it.
In a B-mode ultrasound image, the brightness of each point represents:
Athe depth of the reflector
Bthe frequency shift of the echo
Cthe speed at which the reflector is moving
Dthe strength of the echo from that point
Answer
Answer D. the strength of the echo from that point
B-mode means brightness mode: the grey level codes the echo amplitude, while position in the image codes depth along the beam and the beam direction; frequency shifts are what the Doppler modes display.
A Doppler study measures blood velocity with the beam at 60° to the direction of flow (cos 60° = 0.5). Compared with a beam aligned with the flow, the Doppler shift measured is:
Ahalf as large
Btwice as large
Ca quarter as large
Dthe same
Answer
Answer A. half as large
The Doppler shift Δf = 2f₀v·cos θ / c falls with cos θ, so at 60° it is halved and at 90° it disappears altogether; a small angle to the vessel is therefore sought.
Hardly any echo returns from a boundary between two tissues whose:
Aacoustic impedances are nearly equal
Bdensities differ while the speed of sound is equal
Cthicknesses are equal
Dacoustic impedances differ greatly
Answer
Answer A. acoustic impedances are nearly equal
The share of intensity reflected depends on the mismatch of acoustic impedance, so well-matched tissues send almost nothing back and the beam travels on; differing density alone still changes Z = density × speed and does give an echo.
Extracorporeal shock-wave lithotripsy uses focused ultrasound in order to:
Abreak up a stone in the kidney or the ureter
Bcoagulate a tumour by heating it from outside
Cimage the renal pelvis in real time
Ddissolve a gallstone chemically
Answer
Answer A. break up a stone in the kidney or the ureter
Lithotripsy focuses high-energy shock waves on the stone until it shatters into fragments small enough to pass; it is high-intensity focused ultrasound, not lithotripsy, that coagulates tissue by heating it.
High-intensity focused ultrasound (HIFU) destroys tissue at its focus mainly by:
Aionising the water inside the cells
Bheating it until its proteins coagulate
Cfreezing the tissue until the cells burst
Dreacting with an injected contrast agent
Answer
Answer B. heating it until its proteins coagulate
HIFU concentrates enough acoustic power at the focus to raise the temperature within seconds and coagulate that volume while the tissue the beam crossed stays cool; ultrasound is mechanical, not ionising, radiation.
The axial (depth) resolution of a pulse-echo ultrasound image is set mainly by:
Athe width of the beam at that depth
Bthe pulse repetition frequency
Cthe number of scan lines in the image
Dthe length of the transmitted pulse
Answer
Answer D. the length of the transmitted pulse
Two reflectors lying one beyond the other along the beam are seen separately only if their echoes do not overlap, so a short pulse, which means a high frequency and few cycles, gives the finest axial resolution; the beam width governs lateral resolution.
The energies of the characteristic lines in an X-ray spectrum depend on:
Athe tube current
Bthe exposure time
Cthe anode material
Dthe patient's thickness
Answer
Answer C. the anode material
Characteristic photons come from electrons dropping into vacancies in inner shells (K, L) of the anode atoms, so their energies are set by that element.
Tungsten is used for X-ray tube anodes because it has:
Aa high atomic number and a very high melting point
Ba low atomic number and low density
Cgood transparency to X-rays
Dmagnetic properties
Answer
Answer A. a high atomic number and a very high melting point
Bremsstrahlung output rises with Z (74 for tungsten), and about 99% of the electrons' energy becomes heat, so the target must withstand very high temperatures.
An X-ray beam's intensity falls to one eighth of its original value after 6 mm of a shielding material. The half-value layer of that material for this beam is:
A2 mm
B3 mm
C4.5 mm
D0.75 mm
Answer
Answer A. 2 mm
Attenuation is exponential, I = I₀e^(−µx), so every half-value layer halves the intensity: one eighth is (½)³, three half-value layers in 6 mm, and each is 2 mm thick.
Characteristic X-rays leave the anode of an X-ray tube when:
Aa nucleus in the anode material decays
Ba fast electron is braked in the field of a nucleus
Can outer electron fills a vacancy in an inner shell
Dan electron and a positron annihilate inside the anode
Answer
Answer C. an outer electron fills a vacancy in an inner shell
A bombarding electron ejects a K- or L-shell electron, and the photon released as an outer electron drops into the gap has an energy fixed by the anode element, giving sharp lines; the braking of electrons yields the continuous bremsstrahlung instead.
As the photon energy of a diagnostic X-ray beam is raised, the share of the interactions in soft tissue that are photoelectric:
Afalls, and Compton scattering becomes dominant
Bstays the same, as both depend only on atomic number
Crises, and Compton scattering dies away
Drises, since pair production cannot occur at these energies
Answer
Answer A. falls, and Compton scattering becomes dominant
The photoelectric effect falls away steeply as photon energy rises (and climbs steeply with atomic number), so Compton scattering dominates soft tissue through most of the diagnostic range, costing contrast and adding scatter that fogs the image.
An anti-scatter grid placed between the patient and the X-ray detector works by:
Aremoving low-energy photons from the beam as it leaves the tube
Babsorbing scattered photons before they reach the detector
Climiting the beam to the area of interest
Dturning the X-rays into visible light
Answer
Answer B. absorbing scattered photons before they reach the detector
The lead strips of a grid pass photons that travel straight from the tube and absorb obliquely scattered ones, so contrast is restored; filtration hardens the beam, collimation limits its size and a scintillator makes the light.
In computed tomography, the Hounsfield number of a voxel is calculated from:
Athe number of photons it scatters out of the primary beam
Bthe time the X-rays take to cross it
Cits density compared with that of air
Dits attenuation coefficient compared with that of water
Answer
Answer D. its attenuation coefficient compared with that of water
The Hounsfield scale is defined from the linear attenuation coefficient measured against water, which is fixed at 0 HU, with air at −1000 HU, fat about −100 HU and bone about +1000 HU.
Bspreads the whole range of Hounsfield values over the grey scale
Craises the contrast between tissues of similar density
Dremoves the noise left by the reconstruction
Answer
Answer C. raises the contrast between tissues of similar density
Windowing maps a chosen span of Hounsfield units onto the available grey levels, so a narrow window spreads a small range of attenuation over the whole scale and brings out slight soft-tissue differences; it is a display setting applied after the scan.
Mammography uses a lower tube voltage than chest radiography in order to:
Araise the photoelectric contrast between soft tissues
Bshorten the exposure time that is needed
Creduce the dose absorbed by the glandular breast tissue
Dpenetrate the ribs and the sternum
Answer
Answer A. raise the photoelectric contrast between soft tissues
At 25-32 kV the photoelectric effect still dominates, and because it climbs steeply with atomic number it separates tissues that differ only slightly; the price is a higher absorbed dose, since soft X-rays are absorbed rather than transmitted.
During fluoroscopy, the main source of the radiation reaching staff standing beside the table is:
Athe primary beam from the tube
Bcharacteristic radiation from the detector
Cleakage through the tube housing
Dradiation scattered by the patient
Answer
Answer D. radiation scattered by the patient
The patient acts as the scattering source, so staff dose comes almost entirely from Compton-scattered photons; hence the lead apron, the table-side shield and keeping one's distance, while the primary beam goes into the detector and housing leakage is tiny.
A long TE lets differences in T2 decay show, and a long TR removes T1 effects. T1 weighting uses short TR and short TE; long TR with short TE gives proton density.
Positron emitters for PET such as fluorine-18 are produced in a:
Alinear X-ray tube
Bnuclear generator from molybdenum
Ccyclotron
DMRI magnet
Answer
Answer C. cyclotron
Bombarding targets with protons in a cyclotron makes neutron-poor nuclei, which decay by positron emission. A molybdenum-99 generator supplies technetium-99m for SPECT.
The Larmor frequency of hydrogen is 42.6 MHz per tesla. In a 3 T MRI scanner, hydrogen nuclei therefore precess at about:
A64 MHz
B14 MHz
C128 MHz
D213 MHz
Answer
Answer C. 128 MHz
The Larmor frequency is proportional to the field strength: 42.6 MHz/T × 3 T ≈ 128 MHz, twice the 64 MHz of a 1.5 T magnet, which is why the excitation pulse is a radio-frequency one.
Bthe interval between successive radio-frequency pulses
Cthe decay of the gradient field after a pulse
Dthe decay of the transverse magnetisation
Answer
Answer D. the decay of the transverse magnetisation
T2 measures how fast the spins lose phase coherence so that the transverse signal decays, while T1 is the recovery of longitudinal magnetisation along the main field; the interval between pulses is the repetition time TR.
For any given tissue, the value of T2 compared with T1 is:
Aalways shorter
Balways longer
Calways equal
Dlonger in water and shorter in fat
Answer
Answer A. always shorter
Transverse coherence is lost both by spin-spin interaction and by every process that returns magnetisation to the longitudinal axis, so T2 can never exceed T1; in soft tissue T1 runs to hundreds of milliseconds and T2 to tens.
Clinical MRI images hydrogen nuclei rather than other nuclei because:
Ahydrogen is the only nucleus with a magnetic moment
Bhydrogen carries no electric charge
Chydrogen is the commonest nucleus in body water and fat
Dhydrogen has the longest radioactive half-life of any nucleus
Answer
Answer C. hydrogen is the commonest nucleus in body water and fat
The signal comes from the protons of water and fat, which are present in overwhelming numbers and have a large magnetic moment; nuclei such as ¹³C and ³¹P precess at a Larmor frequency too but are far too scarce for routine imaging.
The radio-frequency power deposited in the patient during an MRI examination is capped chiefly because it:
Aproduces scattered X-rays
Bionises water molecules
Cmagnetises the blood
Dheats the tissue
Answer
Answer D. heats the tissue
Radio waves are non-ionising, and their main biological effect is the tissue heating measured as the specific absorption rate, which the scanner limits; freedom from ionising radiation is a chief advantage of MRI.
The spatial resolution of a clinical PET image, compared with that of CT, is:
Afiner, well under a millimetre
Babout the same as that of CT
Cpoorer, a few millimetres at best
Dfiner than that of MRI
Answer
Answer C. poorer, a few millimetres at best
PET resolution is limited to a few millimetres by the short distance the positron travels before annihilation and by the detector size, so PET is fused with CT, which resolves well under a millimetre, to show where the uptake lies.
Technetium-99m, the nuclide used in most SPECT examinations, emits:
Apositrons that annihilate into two 511 keV photons
Bgamma photons of about 140 keV
Calpha particles of a few MeV
Dbeta particles with a continuous spectrum
Answer
Answer B. gamma photons of about 140 keV
Technetium-99m is a metastable gamma emitter with a half-life of 6 hours and a single 140 keV photon, which suits a gamma camera; positron emitters such as fluorine-18 are what PET, not SPECT, detects.
A PET scanner is usually built into the same gantry as a CT scanner in order to:
Ashorten the half-life of the tracer
Breplace the injected tracer with a contrast medium
Cshow where the tracer uptake lies anatomically
Draise the energy of the annihilation photons
Answer
Answer C. show where the tracer uptake lies anatomically
PET maps metabolism but shows little anatomy, so a CT taken in the same position supplies the anatomical map, and its attenuation data, for the fused PET/CT image.
A vial holds 400 MBq of ¹⁸F-FDG, whose half-life is about 110 minutes. After 5.5 hours the activity left is about:
A100 MBq
B25 MBq
C50 MBq
D133 MBq
Answer
Answer C. 50 MBq
5.5 hours is 330 minutes, that is three half-lives of 110 minutes, so the activity falls by a factor of 2³: 400 MBq / 8 = 50 MBq. Such a short half-life is why FDG must come from a cyclotron near the scanner.
Which is a stochastic effect of ionising radiation?
Askin erythema
Bradiation cataract
Ctemporary sterility
Dradiation-induced leukaemia
Answer
Answer D. radiation-induced leukaemia
Stochastic effects (cancers, heritable effects) have a probability that rises with dose but no threshold; the others are deterministic tissue reactions with a threshold.
In soft tissue, from about 30 keV to about 25 MeV, the dominant photon interaction is:
ACompton scattering
Bthe photoelectric effect
Cpair production
Dcoherent scattering
Answer
Answer A. Compton scattering
Compton scattering depends on electron density, not Z, and dominates in water-like tissue over this range; the photoelectric effect dominates at lower energies.
Alpha emitters are especially dangerous inside the body because alpha particles have:
Aa very long range in tissue
Bno charge
Ca high linear energy transfer, ionising densely along a short track
Da low mass
Answer
Answer C. a high linear energy transfer, ionising densely along a short track
Their energy is deposited within tens of micrometres, causing clustered DNA damage that is hard to repair; outside the body the dead skin layer stops them.
The annual effective dose limit for occupationally exposed workers, averaged over five years, is:
A1 mSv
B20 mSv
C6 mSv
D100 mSv
Answer
Answer B. 20 mSv
Workers may receive up to 20 mSv a year averaged over five years, twenty times the limit of 1 mSv a year set for members of the public; medical exposure of a patient is counted separately.
The world average annual effective dose from natural background radiation is about:
A0.1 mSv
B2.4 mSv
C10 mSv
D50 mSv
Answer
Answer B. 2.4 mSv
Natural sources give about 2.4 mSv a year on world average: radon in the air is the largest part, with cosmic rays, the ground and potassium-40 in the body adding the rest, and local values vary widely.
A radionuclide has a half-life of 10 hours. Taking ln 2 as 0.693, its decay constant is about:
A0.14 per hour
B0.69 per hour
C0.069 per hour
D1.44 per hour
Answer
Answer C. 0.069 per hour
The decay law N = N₀e^(−λt) gives T½ = ln 2 / λ, so λ = 0.693 / 10 h ≈ 0.069 per hour, the fraction of the surviving nuclei decaying each hour; dividing 10 h by 0.693 instead gives 14.4 h, the mean lifetime.
Effective dose differs from equivalent dose in that it also allows for:
Athe time over which the dose was delivered
Bthe type of radiation that delivered the dose
Cthe mass of the tissue that was irradiated
Dhow radiosensitive each irradiated organ is
Answer
Answer D. how radiosensitive each irradiated organ is
Equivalent dose weights the absorbed dose in gray by a radiation weighting factor for the type of radiation; effective dose then weights each organ's equivalent dose by a tissue weighting factor, and both are stated in sievert.
A deterministic effect of ionising radiation, such as skin erythema, is marked out by the fact that:
Ait has no threshold, its probability rising with dose
Bit depends only on the radiation weighting factor used
Cit appears decades later as a malignant tumour
Dit needs a threshold dose, and then worsens with dose
Answer
Answer D. it needs a threshold dose, and then worsens with dose
Deterministic effects need enough dose to kill many cells, so they have a threshold and grow in severity once it is passed; stochastic effects such as radiation-induced cancer are matters of probability with no accepted threshold.
A Geiger-Müller counter registers ionising radiation by means of:
Aa current pulse from an avalanche of ionisation in a gas
Ba change in the resistance of a semiconductor crystal
Cflashes of light in a sodium iodide crystal
Dthe warming of a metal foil placed in the beam
Answer
Answer A. a current pulse from an avalanche of ionisation in a gas
Radiation ionises the gas in the tube and the high voltage multiplies each ion pair into an avalanche, so one countable pulse marks each particle or photon; light flashes in a crystal belong to a scintillation detector instead.
A thermoluminescent dosimeter records the dose a worker has received because, when the crystal is later heated, it:
Achanges colour in proportion to the dose
Bbecomes radioactive in proportion to the dose
Creleases a trapped gas in proportion to the dose
Demits light in proportion to the dose absorbed
Answer
Answer D. emits light in proportion to the dose absorbed
Ionising radiation traps electrons at defects in the crystal, usually lithium fluoride; heating frees them and the light given out is measured to read the accumulated dose, after which the badge can be used again.
The dose rate 1 m from a small gamma source is 80 µSv per hour. At 4 m from the same source it is:
A20 µSv per hour
B10 µSv per hour
C5 µSv per hour
D40 µSv per hour
Answer
Answer C. 5 µSv per hour
By the inverse square law the dose rate falls as 1/r², so four times the distance leaves a sixteenth: 80 / 16 = 5 µSv per hour. Distance is the cheapest of the three protections of time, distance and shielding.
The radiation weighting factor used to turn absorbed dose into equivalent dose is 1 for:
AX-rays, gamma rays and electrons
Balpha particles
Cfast neutrons
Dheavy ions
Answer
Answer A. X-rays, gamma rays and electrons
Photons and electrons ionise sparsely, so they are given a weighting factor of 1 and 1 Gy from them counts as 1 Sv; densely ionising alpha particles carry a factor of 20, and neutrons and heavy ions also exceed 1.
Diathermy and electrosurgery use high-frequency currents (hundreds of kHz and more) because at such frequencies the current:
Aheats tissue without stimulating nerves and muscle
Bstimulates nerves most strongly
Ccannot pass through tissue
Dis completely blocked by the skin
Answer
Answer A. heats tissue without stimulating nerves and muscle
Each half-cycle is too short to depolarise excitable membranes, so the current only heats; at 50 Hz the same current would cause tetany or fibrillation.
The cell membrane acts as an electrical capacitor because it is:
Aa thin insulating lipid layer separating two conducting salt solutions
Ba conducting sheet separating two insulating compartments
Ca semiconductor layer that lets ions pass through it in one direction only
Da magnetic layer that stores energy as magnetic flux
Answer
Answer A. a thin insulating lipid layer separating two conducting salt solutions
Two conductors (the intracellular and extracellular fluid) separated by a thin insulator is exactly a capacitor; the lipid bilayer gives about 1 µF per square centimetre of membrane.
A patch of cell membrane has a capacitance of 1 µF/cm². The charge stored on 1 cm² of it at a membrane potential of 70 mV is about (Q = CV):
A7 nC
B35 nC
C140 nC
D70 nC
Answer
Answer D. 70 nC
Q = CV = 1 × 10⁻⁶ F × 0.070 V = 7 × 10⁻⁸ C, that is 70 nC. The tiny charge needed to polarise the membrane is why only a small fraction of the cell's ions has to move to make an action potential.
Three resistors of 1500 Ω each are connected in parallel. The combined resistance is:
A500 Ω
B1500 Ω
C750 Ω
D4500 Ω
Answer
Answer A. 500 Ω
In parallel the conductances add: 1/R = 3/1500, so R = 500 Ω. Equal resistors in parallel give the resistance of one divided by their number, always less than the smallest; adding them in series would have given 4500 Ω.
A defibrillator stores its shock in a 100 µF capacitor charged to 2000 V. The energy stored is (E = ½CV²):
A50 J
B200 J
C100 J
D400 J
Answer
Answer B. 200 J
E = ½ × 1 × 10⁻⁴ F × (2000 V)² = 200 J, a typical adult defibrillation energy. The energy rises with the square of the voltage, so halving the charging voltage would leave only 50 J.
An electrosurgical generator delivers 0.5 A at 400 V. The power it delivers is (P = VI):
A80 W
B200 W
C400 W
D800 W
Answer
Answer B. 200 W
P = VI = 400 V × 0.5 A = 200 W. This power, concentrated at a small active electrode, is what cuts and coagulates tissue, while the large return electrode spreads the same current over enough skin to stay cool.
Microshock differs from macroshock in that the current:
Apasses between two points on dry, intact skin and is felt as a shock
Breaches the myocardium directly through an intracardiac catheter
Cflows at radiofrequency, so it can only heat the tissue
Dlasts less than one millisecond and so cannot be felt
Answer
Answer B. reaches the myocardium directly through an intracardiac catheter
Microshock bypasses the skin, which carries most of the body's resistance to a low-frequency current, and delivers the whole current to a few grams of myocardium, so a current about a thousand times smaller than the roughly 100 mA needed at the skin can cause ventricular fibrillation.
A 20 kΩ resistance in series with a 5 µF capacitor has a time constant (τ = RC) of:
A0.02 s
B0.5 s
C0.1 s
D2 s
Answer
Answer C. 0.1 s
τ = RC = 2 × 10⁴ Ω × 5 × 10⁻⁶ F = 0.1 s. The time constant is the time to reach about 63% of the final voltage, and it sets how fast an electrode-and-amplifier combination can follow a changing biopotential.
Transcranial magnetic stimulation makes cortical neurones fire because the rapidly changing magnetic field of the coil:
Aaligns the magnetic moments of the neuronal membranes
Binduces an electric field, and so a current, in the tissue under it
Cheats the cortex by raising its specific absorption rate
Dexerts a direct magnetic force on the ions held inside the cell body
Answer
Answer B. induces an electric field, and so a current, in the tissue under it
By Faraday's law a changing magnetic field induces currents in a conductor, and those induced currents depolarise the neurones; a static field of the same strength, as in an MRI scanner, induces nothing and does not stimulate nerves.
A charged particle moving through a uniform magnetic field feels the largest force when its velocity is:
Aparallel to the field
Bantiparallel to the field
Cperpendicular to the field
Dat 45° to the field
Answer
Answer C. perpendicular to the field
The force is F = qvB sin θ, greatest when θ = 90° and zero when the particle moves along the field. The force is also always perpendicular to the velocity, so a magnetic field in tesla bends a charged particle's path without changing its speed.
A systolic blood pressure of 120 mmHg equals about (1 mmHg = 133.3 Pa):
A12 kPa
B101 kPa
C24 kPa
D16 kPa
Answer
Answer D. 16 kPa
120 × 133.3 Pa = about 16 000 Pa, that is 16 kPa. The pascal is the SI unit, but clinical pressures are still quoted in millimetres of mercury; 101 kPa is atmospheric pressure, against which arterial pressure is measured.
A bag of saline (density 1000 kg/m³) hangs with its fluid level 1 m higher than the needle in the vein. The hydrostatic pressure it adds at the needle is about (P = ρgh, g = 9.81 m/s², 1 mmHg = 133.3 Pa):
A7 mmHg
B15 mmHg
C37 mmHg
D74 mmHg
Answer
Answer D. 74 mmHg
P = ρgh = 1000 × 9.81 × 1 = 9810 Pa, and 9810/133.3 is about 74 mmHg, comfortably more than venous pressure, so the drip runs in. Raising the bag higher raises the pressure in direct proportion to the height.
The energy of a photon of frequency 6 × 10¹⁴ Hz is about (E = hf, h = 6.63 × 10⁻³⁴ J·s):
A0.8 × 10⁻¹⁹ J
B4.0 × 10⁻¹⁹ J
C2.0 × 10⁻¹⁹ J
D1.3 × 10⁻¹⁹ J
Answer
Answer B. 4.0 × 10⁻¹⁹ J
E = hf = 6.63 × 10⁻³⁴ × 6 × 10¹⁴ = about 4.0 × 10⁻¹⁹ J, a visible photon of roughly 2.5 eV. Photon energy rises with frequency, which is why ultraviolet light damages molecules and infrared only warms them.
A pulse oximeter measures arterial oxygen saturation by comparing the absorption of:
Aultraviolet and blue light in the capillary bed
Bred and infrared light in the pulsating arterial part of the signal
Ctwo ultrasound frequencies reflected from the red cells in the finger
Dinfrared light before and after the finger is warmed
Answer
Answer B. red and infrared light in the pulsating arterial part of the signal
Oxyhaemoglobin and deoxyhaemoglobin absorb red (about 660 nm) and infrared (about 940 nm) light by different amounts, and taking only the pulsatile component removes the absorption of skin, bone and venous blood.
Taking the latent heat of vaporisation of water at body temperature as 2.4 MJ/kg, evaporating 100 g of sweat removes about:
A12 kJ
B240 kJ
C120 kJ
D24 kJ
Answer
Answer B. 240 kJ
0.1 kg × 2.4 × 10⁶ J/kg = 2.4 × 10⁵ J, that is about 240 kJ. Evaporation is the only route that still works when the surroundings are hotter than the skin, which is why sweat that drips off unevaporated cools nothing.
In a laboratory centrifuge, the relative centrifugal force on the sample is proportional to:
Athe angular speed
Bthe square root of the angular speed
Cthe square of the angular speed
Dthe cube of the angular speed
Answer
Answer C. the square of the angular speed
The centripetal acceleration is ω²r, so the relative centrifugal force (that acceleration divided by g) rises with the square of the angular speed: doubling the revolutions per minute quadruples it, while doubling the rotor radius only doubles it.
Thermography images the body by detecting radiation emitted by the skin, which at body temperature lies mainly in the:
Aultraviolet
Bvisible red
Cinfrared
Dmicrowave
Answer
Answer C. infrared
A surface near 33 °C radiates most strongly at a wavelength of about 9-10 µm, in the infrared; by the Stefan-Boltzmann law the power radiated rises steeply with absolute temperature, so an inflamed area shows as a hot patch.
Phototherapy for neonatal jaundice uses blue light because the light:
Ais absorbed by bilirubin and converts it into water-soluble isomers
Bwarms the skin so that bilirubin is broken down by the heat
Cdestroys the gut bacteria that release bilirubin from bile
Dmakes the skin produce vitamin D, which carries the bilirubin away in bile
Answer
Answer A. is absorbed by bilirubin and converts it into water-soluble isomers
Bilirubin absorbs strongly in the blue (about 450 nm); the absorbed photons isomerise it into forms that the newborn can excrete in bile and urine without the conjugation its liver cannot yet manage.
By the Beer-Lambert law, the intensity of light passing through a uniform absorbing solution falls:
Alinearly with the path length
Bas the inverse square of the path length
Cas the square of the path length
Dexponentially with the path length
Answer
Answer D. exponentially with the path length
Each thin layer absorbs the same fraction of the light reaching it, so the intensity decays exponentially: I = I₀ × 10⁻ᵋᶜˡ. That is why absorbance, the logarithm of the intensity ratio, is the quantity that is proportional to concentration and path length.
The same charge passes through a silver cell (Ag⁺, M = 108 g/mol) and a copper cell (Cu²⁺, M = 63.5 g/mol) in series. The mass of silver deposited is about:
Athe same as the copper
B1.7 times the copper
C3.4 times the copper
Dhalf the copper
Answer
Answer C. 3.4 times the copper
By Faraday's second law, mass goes as M/z: 108/1 = 108 for silver against 63.5/2 ≈ 31.8 for copper, a ratio of about 3.4.
What mass of silver does a charge of 1930 C deposit from a solution of Ag⁺ ions (M = 108 g/mol, F ≈ 96 500 C/mol)?
A0.54 g
B1.08 g
C4.32 g
D2.16 g
Answer
Answer D. 2.16 g
1930 C is 1930/96 500 = 0.02 mol of electrons, and silver is singly charged, so 0.02 mol of silver is plated: 0.02 × 108 = 2.16 g. With a divalent ion such as Cu²⁺ the same charge would deposit only 0.01 mol.
A plating cell is to deposit 0.05 mol of copper from Cu²⁺ ions at a steady current of 5 A (F ≈ 96 500 C/mol). The time needed is about:
A16 min
B32 min
C64 min
D97 min
Answer
Answer B. 32 min
Each mole of Cu²⁺ needs 2 mol of electrons, so the charge is 0.05 × 2 × 96 500 = 9650 C; at 5 A that takes 9650/5 = 1930 s, about 32 min. Forgetting the factor z = 2 would halve the answer.
An electrolytic cell differs from a galvanic cell in that in the electrolytic cell:
Athe current through the solution is carried by electrons rather than ions
Bthe reaction runs spontaneously and delivers electrical energy
Cboth electrodes must be made of the same metal
Dan external supply drives a reaction that would not happen on its own
Answer
Answer D. an external supply drives a reaction that would not happen on its own
Electrolysis uses electrical energy to force a non-spontaneous change, such as plating copper onto the cathode; a galvanic cell is the reverse, a spontaneous reaction that produces a potential difference. In either cell, the solution carries current as moving ions.
In any electrolytic cell, the half-reaction taking place at the anode is:
Areduction, the gain of electrons
Boxidation, the loss of electrons
Cneutralisation of the electrolyte
Dprecipitation of the dissolved salt
Answer
Answer B. oxidation, the loss of electrons
Oxidation happens at the anode and reduction at the cathode in every cell; in the copper electrolysis lab this is why copper dissolves from the anode as Cu²⁺ while the mass of the cathode grows by the same amount.
The Nernst equation gives the equilibrium potential of a membrane for one permeant ion from:
Athe total ionic strength of the two solutions
Bthe difference between those two concentrations, in mmol/L
Cthe ratio of that ion's concentrations on the two sides
Dthe membrane's capacitance per unit area
Answer
Answer C. the ratio of that ion's concentrations on the two sides
The potential is (RT/zF) times the natural logarithm of the concentration ratio, with F the Faraday constant, so only the ratio matters: the same tenfold ratio gives the same potential whether the concentrations are high or low.
Taking the Nernst slope at 37 °C as 61 mV per tenfold ratio for a singly charged ion, the equilibrium potential for potassium at 140 mmol/L inside and 14 mmol/L outside, measured inside against outside, is about:
A−61 mV
B−94 mV
C−122 mV
D−30 mV
Answer
Answer A. −61 mV
E = 61 × log₁₀(14/140) = 61 × (−1) = about −61 mV. The inside is negative because potassium leaving down its tenfold gradient leaves unbalanced negative charge inside; a hundredfold gradient would give about −122 mV.
Iontophoresis delivers a charged drug through the skin by:
Aapplying a direct current so the drug ions migrate to the opposite electrode
Bheating the skin with a high-frequency current so the drug diffuses faster
Cdriving the drug in with brief pulses of focused ultrasound
Dchanging the skin's pH so that the drug becomes uncharged
Answer
Answer A. applying a direct current so the drug ions migrate to the opposite electrode
A small direct current makes the charged drug move through the skin like any ion in an electrolysis cell, repelled from the electrode of its own sign; the dose delivered is proportional to the charge passed, as in Faraday's first law.
A direct current left running through a skin electrode for a long time can cause a chemical (galvanic) burn because at the electrode the current:
Ainduces eddy currents that tear the deeper tissue
Bturns entirely into heat and takes the skin to more than 60 °C within one second
Cmagnetises the electrode so that it sticks to the skin
Delectrolyses tissue fluid, leaving acid at the anode and alkali at the cathode
Answer
Answer D. electrolyses tissue fluid, leaving acid at the anode and alkali at the cathode
Direct current must be carried across the electrode-tissue boundary by a chemical reaction: water is oxidised at the anode, releasing H⁺, and reduced at the cathode, releasing OH⁻, so the tissue is damaged by pH extremes. Alternating currents reverse before much product builds up.
The eye's optics (cornea and lens, about 60 dioptres) and the packing of cones in the fovea of the retina set this limit, measured in the lab with lines on a chart.
Light inside a medium of refractive index 1.5 meets a boundary with air (n = 1). The critical angle for total internal reflection is about (sin θc = 1/1.5):
A24°
B34°
C42°
D49°
Answer
Answer C. 42°
sin θc = 1/1.5 = 0.667, so θc is about 42°; light striking the boundary at more than this angle is reflected entirely back into the glass, which is how the fibres of an endoscope carry light round bends.
Light passing from air into the cornea (n ≈ 1.38) at an oblique angle:
Abends away from the normal and travels faster
Bbends away from the normal and travels more slowly
Cbends toward the normal and travels faster
Dbends toward the normal and travels more slowly
Answer
Answer D. bends toward the normal and travels more slowly
By Snell's law, n₁ sin θ₁ = n₂ sin θ₂, so entering the denser medium makes the angle to the normal smaller, and the speed falls to c/n. This refraction at the air-cornea surface supplies about two thirds of the eye's roughly 60 dioptres.
A patient's far point is 50 cm from the eye. The spectacle lens that brings distant objects into focus is:
Aa diverging lens of 2 D
Ba converging lens of 2 D
Ca converging lens of 4 D
Da diverging lens of 4 D
Answer
Answer A. a diverging lens of 2 D
The lens must make parallel rays appear to come from the far point, so its focal length is −0.50 m and its power 1/(−0.5) = −2 D. A diverging lens is needed because the myopic eye is already too powerful for its length.
Accommodation for near vision in the human eye is produced by:
Aa change in the curvature of the cornea
Bthe ciliary muscle relaxing, so that the lens flattens
Cthe ciliary muscle contracting, letting the lens become more curved
Dlengthening of the whole eyeball, which shifts the position of the retina
Answer
Answer C. the ciliary muscle contracting, letting the lens become more curved
Contraction of the ciliary muscle slackens the suspensory fibres, so the elastic lens rounds up and adds dioptres to the eye's power. The cornea's power is fixed; presbyopia is the loss of this lens change with age, corrected with a converging reading lens.
A patient reads at 6 m only the Snellen line that a person with normal sight reads at 24 m. The visual acuity is recorded as:
A4/6
B24/6
C6/12
D6/24
Answer
Answer D. 6/24
Snellen acuity is written as the test distance over the distance at which a normal eye could read the smallest line seen, so 6/24 here; the larger the lower number, the worse the acuity, and 6/6 is normal.
Phase-contrast microscopy makes unstained living cells visible by turning:
Adifferences in fluorescence into differences in colour
Bthe polarisation of the light into interference colours
Cdifferences in optical path into differences in brightness
Ddifferences in electron density into differences in contrast
Answer
Answer C. differences in optical path into differences in brightness
Parts of the cell differ in refractive index and thickness, so they shift the phase of the light without absorbing it; a phase plate in the objective converts those phase shifts into amplitude differences the eye can see, with no staining or fixation.
Magnifying a light-microscope image beyond about 1000× gives no further detail (empty magnification) because the resolution is set by:
Athe focal length of the eyepiece
Bthe brightness of the condenser lamp
Cthe wavelength of the light and the numerical aperture
Dthe thickness of the specimen mounted on the glass slide
Answer
Answer C. the wavelength of the light and the numerical aperture
The smallest resolvable separation is about 0.61λ/NA, so with visible light and an oil-immersion NA of about 1.4 it is around 0.2 µm; enlarging further only makes a blurred image bigger.
Chromatic aberration in a simple lens arises because the refractive index of the glass is:
Aslightly greater for blue light than for red
Bslightly greater for red light than for blue
Cthe same for all colours, while the surfaces are not spherical
Ddependent on how brightly the lens is lit
Answer
Answer A. slightly greater for blue light than for red
Dispersion makes the shorter blue wavelengths refract more, so they focus closer to the lens than the red, fringing the image with colour; an achromatic doublet of two glasses with opposite dispersion cancels most of it.
A sound has a loudness of n phons when it sounds as loud as a 1 kHz tone of n dB; equal-loudness contours show that the ear is less sensitive at low and very high frequencies.
The main job of the middle ear ossicles and tympanic membrane is to:
Amatch the impedance of air to that of the cochlear fluid
Bfilter out low frequencies
Cconvert sound into nerve impulses
Damplify the frequency
Answer
Answer A. match the impedance of air to that of the cochlear fluid
The large tympanic membrane funnels force onto the small oval window and the ossicles act as a lever, so most sound energy is transmitted rather than reflected.
The ear canal is about 2.5 cm long and closed at the tympanic membrane, so it behaves as a quarter-wave resonator. With sound travelling at 343 m/s, its resonant frequency is about:
A3.4 kHz
B1.7 kHz
C850 Hz
D6.8 kHz
Answer
Answer A. 3.4 kHz
A closed pipe resonates when its length is a quarter of the wavelength, so λ = 4 × 0.025 = 0.1 m and f = 343/0.1 = about 3.4 kHz. This resonance helps make the ear most sensitive at about 2-5 kHz, the band that carries speech.
Sound A has a level 20 dB higher than sound B. Its intensity is greater by a factor of:
A2
B100
C50
D20
Answer
Answer B. 100
The level is 10 log(I/I₀), so 20 dB means log(I_A/I_B) = 2 and the intensity ratio is 100. Each 10 dB is a further tenfold step in intensity, while 3 dB is only a doubling.
A small source in the open radiates sound equally in all directions and gives 70 dB at 1 m. At 10 m the level is about:
A35 dB
B60 dB
C50 dB
D67 dB
Answer
Answer C. 50 dB
Intensity falls as the inverse square of the distance, so ten times the distance is a hundredth of the intensity, and a hundredfold drop in intensity is 10 log(1/100) = −20 dB, giving 50 dB.
In a Weber test on a patient with a conductive hearing loss of the right ear, a tuning fork placed on the middle of the forehead is heard:
Alouder in the right ear
Blouder in the left ear
Cequally loudly in both ears
Din neither ear
Answer
Answer A. louder in the right ear
The sound reaches both cochleae by bone conduction, which is normal in conductive loss, and in the blocked right ear less of that bone-conducted sound escapes through the middle ear and less room noise masks it, so the tone lateralises to the affected side. In sensorineural loss it lateralises to the better ear instead.
Years of work in loud industrial noise typically show on the audiogram as:
Aa flat loss of the same size at every frequency
Ba loss confined to frequencies under 500 Hz
Ca notch near 4 kHz, with better thresholds at 8 kHz
Dan air-bone gap of about 30 dB at every frequency
Answer
Answer C. a notch near 4 kHz, with better thresholds at 8 kHz
Noise damages the cochlea first in the region tuned near 4 kHz, so the audiogram dips there and recovers somewhat at 8 kHz; an air-bone gap would instead point to a conductive problem, and presbycusis slopes down steadily toward the highest frequencies.
An ambulance siren of fixed frequency is heard by a bystander as the ambulance drives toward them. The siren is heard at a:
Afrequency equal to the siren's, but a longer wavelength
Blower frequency than the siren emits
Chigher frequency than the siren emits
Dfrequency equal to the siren's, but a lower intensity
Answer
Answer C. higher frequency than the siren emits
Approaching motion crowds the wavefronts together, shortening the wavelength and raising the frequency at the listener (the Doppler effect); the pitch drops abruptly as the ambulance passes and begins to recede.
A vibration of 10 Hz, which the ear cannot hear as a tone, is classed as:
Aultrasound
Binfrasound
Can audible pure tone
Da supersonic wave
Answer
Answer B. infrasound
Hearing spans about 20 Hz to 20 kHz, so anything under 20 Hz is infrasound and anything over 20 kHz is ultrasound; infrasound is felt as vibration rather than heard, which is why very low frequencies never appear on a standard audiogram.
The middle ear raises the pressure delivered to the cochlear fluid mainly because:
Athe ossicles vibrate at a higher frequency than the tympanic membrane does
Bthe round window moves in step with the stapes
Cthe cochlear fluid is less dense than air
Dthe tympanic membrane has a far larger area than the stapes footplate
Answer
Answer D. the tympanic membrane has a far larger area than the stapes footplate
The same force collected over the wide tympanic membrane is concentrated on the small stapes footplate, and the ossicular lever adds a little more, which is how the middle ear matches the impedance of air to that of the much stiffer cochlear fluid.
The bell of a stethoscope, held lightly against the chest, transmits best the:
Alow frequencies, while the diaphragm favours the high ones
Bhigh frequencies, but only when it is pressed down firmly
Csame frequencies as the diaphragm, but more loudly
Dhigh frequencies, while the diaphragm favours the low ones
Answer
Answer A. low frequencies, while the diaphragm favours the high ones
The slack skin inside the lightly applied bell moves easily and passes low-pitched sounds; pressing the bell down stretches the skin into a diaphragm and cuts them out, which is exactly how the stiff diaphragm side selects the high frequencies.
The QRS is equiphasic (isoelectric) in aVL and most positive in lead II. The electrical axis is about:
A+60°
B−30°
C+150°
D0°
Answer
Answer A. +60°
The axis is perpendicular to the lead in which QRS is equiphasic: −30° ± 90° gives +60° or −120°, and the large positive QRS in lead II (+60°) picks +60°.
The P wave of the electrocardiogram is produced by:
Adepolarisation of the atria
Brepolarisation of the atria
Cdepolarisation of the ventricles
Dconduction of the impulse through the atrioventricular node
Answer
Answer A. depolarisation of the atria
The P wave is the spread of depolarisation over the atrial muscle; the atrioventricular delay that follows it produces no deflection and is seen as the flat PR segment, and the much larger ventricular depolarisation is the QRS complex.
On an electrocardiogram run at 25 mm/s, two neighbouring R waves are 20 mm apart. The heart rate is:
A60/min
B80/min
C75/min
D100/min
Answer
Answer C. 75/min
20 mm at 25 mm/s is 0.8 s per beat, so the rate is 60/0.8 = 75/min. The same result follows from the grid: 20 mm is 4 large squares, and 300/4 = 75/min.
The ST segment of the electrocardiogram corresponds to the period when:
Athe atria are repolarising
Bthe sinoatrial node is depolarising
Cthe ventricles are filling before atrial contraction
Dthe whole ventricular myocardium is depolarised
Answer
Answer D. the whole ventricular myocardium is depolarised
Between the end of the QRS complex and the start of the T wave every ventricular cell is depolarised, so there is no potential difference to record and the segment is normally level with the baseline; injury to part of the myocardium breaks that uniformity and shifts it.
Hyperkalaemia characteristically changes the electrocardiogram by producing:
Atall, peaked T waves
BT waves inverted in every lead
Cprominent U waves after the T wave
Da short PR interval with a slurred QRS upstroke
Answer
Answer A. tall, peaked T waves
A raised extracellular potassium speeds ventricular repolarisation, so the T wave becomes tall, narrow and peaked, often with a widening QRS as the level climbs; prominent U waves are the hallmark of a low potassium instead.
A regular 50 Hz sinusoidal interference running through the whole electrocardiogram trace comes from:
Amovement of the chest wall with each breath
Btremor of the patient's shoulder muscles
Can electrode that has dried out and begun to lift off the skin
Dthe mains supply, picked up by the patient and the leads
Answer
Answer D. the mains supply, picked up by the patient and the leads
Mains hum enters as 50 Hz (or 60 Hz) and is removed with a notch filter or by checking earthing and lead contact; muscle tremor gives irregular high-frequency fuzz, and breathing or a loose electrode makes the baseline wander slowly.
The right-leg electrode of a standard 12-lead electrocardiogram serves as:
Aa reference (earth) electrode that enters no lead
Bthe negative pole of lead aVR
Cthe positive pole of lead III
Dthe common positive pole shared by the six chest leads
Answer
Answer A. a reference (earth) electrode that enters no lead
Only the right arm, left arm and left leg form Einthoven's triangle and its augmented leads; the right-leg electrode is a reference that is driven to cancel interference, so a poor contact there adds noise but changes no deflection.
Arecord a 12-lead electrocardiogram at the fastest available paper speed
Brecord the electrocardiogram during treadmill exercise to a target rate
Crecord the electrocardiogram over 24 hours or more of ordinary activity
Dmeasure the electrical axis more precisely than a resting recording can
Answer
Answer C. record the electrocardiogram over 24 hours or more of ordinary activity
A resting trace lasts only seconds, so intermittent arrhythmias and paroxysmal symptoms are missed; the Holter recorder stores a continuous trace through a normal day and night, with a diary to match symptoms to the rhythm. Exercise to a target rate is a stress test instead.
The chest leads V1-V6 differ from limb leads I, II and III in being:
Abipolar, each recorded between two chest electrodes
Bunipolar, each recorded against a common reference
Crecorded at twice the paper speed of the limb leads
Drecorded in the frontal plane rather than the horizontal one
Answer
Answer B. unipolar, each recorded against a common reference
Each precordial electrode is compared with Wilson's central terminal, the average of the three limb electrodes, whereas leads I, II and III each measure the potential difference between two limbs of Einthoven's triangle. The chest leads view the heart in the horizontal plane.
Atrial fibrillation appears on the electrocardiogram as:
Aregular atrial waves near 300/min with a sawtooth baseline
Ba regular rhythm with one P wave in front of every QRS complex
Can irregularly irregular ventricular rhythm with no P waves
Dwide QRS complexes at a regular rate of about 150/min
Answer
Answer C. an irregularly irregular ventricular rhythm with no P waves
Chaotic atrial activity replaces the P wave with a fine wavy baseline, and the atrioventricular node passes impulses erratically, so the R-R intervals vary from beat to beat. Sawtooth waves near 300/min are atrial flutter, and wide regular complexes suggest a ventricular rhythm.
In an irregular rhythm such as atrial fibrillation, the ventricular rate is best estimated by:
Acounting the P waves in a 6-second strip and multiplying that count by 10
Bdividing 300 by the shortest R-R interval in large squares
Cdividing 1500 by the longest R-R interval in small squares
Dcounting the QRS complexes in a 6-second strip and multiplying by 10
Answer
Answer D. counting the QRS complexes in a 6-second strip and multiplying by 10
The 300-over-large-squares rule assumes every R-R interval is the same, so in an irregular rhythm it must be replaced by an average: at 25 mm/s a 6-second strip is 30 large squares, and the complexes counted in it times ten give the rate per minute. Atrial fibrillation has no P waves to count.